Inconsistent Results between EGSnrc v4 (rev. 2.2.3) and EGSnrc v2021 #772
Unanswered
agsponer-metas
asked this question in
Questions
Replies: 1 comment 1 reply
-
@agsponer-metas, thank you for taking the time to report this in great detail. I will promote your message to an "Issue", as there is enough in here indeed to investigate further. The material you provide here will prove extremely useful. @blakewalters, can you take a peek at this to see if it rings a bell? 🔔 We have the source code versions going back EGSnrc v4 rev. 2.2.3, so I will at my end look for any obvious changes or commits relating to what you oberved, and let you know. |
Beta Was this translation helpful? Give feedback.
1 reply
Sign up for free
to join this conversation on GitHub.
Already have an account?
Sign in to comment
-
I'm new to EGSnrc and have been recently tasked with "resurrecting" an old BEAMnrc simulation of a treatment head attached to an electron accelerator. At the time the old simulations were performed (~2003-2006), a cluster of office type machines was used to run the simulations. In order to retire this old cluster and because we want to run new simulations which modify the existing geometry of the treatment head, it was decided to migrate the simulations to new hardware and the new version of EGSnrc (v2021). The old simulations were performed using EGSnrc v4 rev. 2.2.3. (which I will call v4 for brevity) and BEAMnrc Rev. 1.78.
To verify that the results obtained by EGSnrc v2021 are unchangend compared to the old results (EGSnrc v4), I ran the simulations using EGSnrc v2021 and the same code that was used for the older runs (same
.egsinp
files, BEAM accelerator modules, PEGS4 files). For electron beams this worked fine and I was able to reproduce the exact same results as EGSnrc v4 (inside the statistical uncertainty).However, for photon beams (which use a gold conversion target) a statistically significant difference in the energy spectrum of primary and secondary electrons (which reach the scoring plane after the first part of the treatment head) was observed, see
EGSnrc_v4(2.2.3)_vs_v2021_for_entire_treatment_head.pdf
in the attachments. In the histogram, the particle count has been normalized to the incident electrons originating from the accelerator "source" and to the area of the scoring plane (33 x 33 cm^2).The rest energy of the electron has been subtracted for the electrons/positrons and the error bars are assumed to be sqrt(N), where N is the amount of particles in a bin. In the v2021 version of EGSnrc, around 20% more primary electrons reach the scoring plane. This is significant for us, for example when we want to characterize the electron contamination in the photon beam.
In order to try and narrow down if this difference stems from a single component module or is the result from interactions along the entire length of the treatment head, I ran simulations using subsets of the treatment head geometry component modules. If I just consider the first two component modules (the target in its holder
targ_hol
and the cone behind ittar_cone
), the results from EGSnrc v4 and v2021 seem to match. However, if the next CM (a copper cooling SLABcool_cap
) is added the results start to diverge (see "EGSnrc_v4(2.2.3)_vs_v2021.pdf").For this situation, I provided a minimum reproducible example (BEAM_cool_cap_test in the attachments). Looking at the
.egslst
output files (also attached) for this third component module, I see a difference in the geometry:For some reason, EGSnrc v4 inserts the
cool_cap
SLAB at a position of 2.66 cm, even thoughZ_MIN
is specified as 2.67 in the.egsinp
file (see above). This means that the SLAB is flush with the CONS3R module in front of it, which extends to 2.66 cm. EGSnrc v2021, however, correctly inserts an air gap of 0.01 cm and puts the SLAB at a position of 2.67 cm.The interesting thing is now, if I increase the thickness of the copper
cool_cap
SLAB by 0.01 cm in EGSnrc v2021 (see below and the energy spectrum inEGSnrc_v4(2.2.3)_vs_EGSnrc_v2021+0.01_cm_Cu.pdf
), I can reproduce the same energy spectrum as EGSnrc v4!Does this mean that EGSnrc v4 (rev. 2.2.3) erroneously inserts 0.01 cm of additional material? Is this a bug in EGSnrc v4 which has been fixed at some point between v4 and v2021?
Adding the 0.01 cm of copper does not fix the issue for the simulation of the entire treatment head, as there seems to be a similar error occurring in a second CM later on (which I could not pinpoint yet). I will have to go from component module to component module to find out which of them introduces further errors. Compared to the
cool_cap
SLAB, however, I have not been able to find any further discrepancies in the.egslst
log files.As I have around 15 simulations (with different electron beam energies and flattening filters) to go through, I would be glad for any aid in explaining these differences between EGSnrc v4 (rev. 2.2.3) and v2021. Was there maybe some change to the geometry generation algorithm? Would this imply that the results using the old version of EGSnrc are incorrect?
I have not yet played around with the Monte Carlo transport parameters (where some default values between the EGSnrc versions might have changed which I do not explicitly define in the
.egsinp
file), but from the electron beam simulations (without the conversion target) I did not yet find any transport parameters which lead to significant differences.I hope that the information I provided is complete and clear enough to describe my issue.
Thanks in advance for your reply,
Andreas
attachments.zip
Beta Was this translation helpful? Give feedback.